How to prepare samples for equilibrium SAXS
SAXS sample shreadsheets
BioCAT requests that for each beamtime you send us a spreadsheet of your samples and buffers at least a week before the beamtime starts or you ship your samples (whichever is earlier). Please note that there are two tabs, one for sample and one for buffer. You may also find it useful for tracking/planning your experiments. If you’re not sure what the final concentrations are a week out, send us the spreadsheet with target concentrations, then send us an update once you’ve finalized the preparation. This helps us make sure the appropriate equipment is available for your beamtime, and also lets us recommend any changes that might be needed to sample or buffer volumes/quantities/concentrations ahead of time, ensuring a more successful beamtime.
Checklist for SAXS sample preparation
Below are the recommendations for buffer and sample preparation for SAXS. Note that at the end of the day, these are guidelines and the goal is always the happiness and stability of the sample. We can accommodate most buffers, so ask your beamline scientist if you have an unusual buffer composition.
Also, if possible we recommend combining similar buffers, particularly for separation-coupled measurements like SEC-SAXS. Fewer buffers will save you a significant amount of time during your experiment, so consider whether you can combine similar conditions.
Buffer component recommendations
  • 20-100 mM PBS, Tris, HEPES, etc. PBS is preferred if possible.
  • Avoid precipitation-prone compositions (such as calcium salts in phosphate buffers).
  • Salt concentration < 1M, typically 50-200 mM.
  • If you need a reductant, use longer lived ones like TCEP rather than DTT (especially with higher pH buffers).
  • As the limiting factor for radiation damage is typically sheath buffer composition as opposed to the macromolecular sample itself, additives often used specifically to protect against damage (such as glycerol) are not recommended.
  • As some additives, such as TCEP, can change buffer pH; it is recommended to pH-adjust your buffer after addition, or to pH adjust any additive stocks.
Coflow sheath buffers
  • BioCAT’s SAXS sample cell utilizes a coflow sheath design, where the central (sample) stream is protected from the capillary walls by a fluidic sheath. This necessitates a separate buffer in addition to the sample buffer.
  • As sheath buffer composition is typically our limiting factor in terms of radiation damage, if your running buffer/sample buffer contains radical scavengers (such as reductants, or even some buffers themselves such as Tris or HEPES), it is recommended that you provide a separate coflow sheath buffer. If possible, phosphate buffers are preferred.
  • It is recommended that the sheath buffer be pH-matched to the sample buffer, as some pH equilibration can take place between the sample stream and coflow sheath. Matching the ionic strength, or at least the salt concentration, can also be helpful for contrast.
  • For example: for an SEC running buffer compsition of 20 mM Tris, 100 mM NaCl, 1 mM TCEP pH 7.4, the recommended sheath buffer composition would be 20 mM sodium or potassium phosphate, 100 mM NaCl pH 7.4.
Buffer matching
  • SEC-SAXS/SEC-MALS-SAXS - Best if buffer is perfectly matched, but can tolerate nominally matched buffers that are not matched through exchange.
  • Batch SAXS - Must be a perfect match. Use dialysis buffer, the buffer from the final SEC purification step, or the flow through from the final concentration step.
  • For batch mode, it is recommended that additives such as substrates also be prepared in the perfectly matched sample buffer if possible.
If mailing concentrated buffer for separation coupled measurements (e.g. a 10x stock for use with SEC-SAXS), we recommend you prepare the concentrated solution, create the final dilution in your lab, exchange your samples into that, and then mail us the remaining concentrated stock for dilution on site. While this doesn’t perfectly match buffer, it gets close, and the buffer exchange on the column will provide the necessary perfect match.
Buffer volume
Sample (Protein or DNA/RNA) concentration and volume
  • Concentration:
    • SEC-MALS-SAXS/SEC-SAXS: 240/(MW in kDa) = concentration mg/ml
    • Batch mode SAXS: 60/(MW in kDa) = concentration in mg/ml. Use a dilution series with a least 2 other concentrations bracketed around this. E.g. for a 20 kDa protein, measure at 6, 3, and 1.5 mg/ml.
For nucleic acid samples, you can divide the above concentrations by 2.5.
  • Minimum volume:
    • SEC-SAXS/SEC-MALS-SAXS - 250 µL
    • Batch-mode SAXS - 10 µL per concentration
Note: If mail-in samples are shipped frozen, they will be fast thawed (e.g. in hand/water bath) unless you specify that the samples should be thawed on ice.
Quality control: Pre-examination before X-ray scattering (Monodispersity)
  • Gel filtration (single and symmetric peak)
  • SDS-PAGE or Native GEL (single band)
  • Centrifugation or filter (> 15,000 x g * 5-10 mins)
  • Dynamic light scattering
  • SEC-MALS
  • Others
More details
In the following video from the 2019 Everything BioSAXS 5 workshop, Dr. Kushol Gupta discusses what you need to know about sample preparation for SAXS (get slides)